ISO standard 2720 gives requirements for the behavior of photoelectric photographic exposure meters, both reflected light and incident light types. It does this by way of equations for each type of meter that include a "calibration constant", symbolized as K for reflected light meters and C for incident light meters.
But no values are prescribed (or even suggested) for these constants, only for each one an "acceptable" range (fairly wide). This recognizes the reality that this standard is meant to "codify" (but not normalize) the diverse existing practices of exposure meter manufactures as to the "calibration" of their meters. Thus a meter manufacturer could adopt that value of K (or C) for their meter they have found gives results that suit most of the users most of the time.
The standard then says: "The criterion of overall accuracy of the meter shall be the deviation, from the stated value, of the constant K [or C, as applicable]."
The standard then gives a quite detailed procedure for determining the actual value of K [or C] exhibited by a meter under test.
Yet nowhere does the standard express (or even suggest) a degree of the deviation of the value of K [or C] from the stated value that is "permitted" under the standard.
So for some meter model, the manufacturer may state that it uses a value of 12.5 for K. In a laboratory, we determine by scrupulous and tedious testing the actual value of K practiced by a specimen meter and find that to be 11.2. Does that meter conform to the standard as to its accuracy? That is not defined.
Best regards,
Doug
But no values are prescribed (or even suggested) for these constants, only for each one an "acceptable" range (fairly wide). This recognizes the reality that this standard is meant to "codify" (but not normalize) the diverse existing practices of exposure meter manufactures as to the "calibration" of their meters. Thus a meter manufacturer could adopt that value of K (or C) for their meter they have found gives results that suit most of the users most of the time.
The standard then says: "The criterion of overall accuracy of the meter shall be the deviation, from the stated value, of the constant K [or C, as applicable]."
The standard then gives a quite detailed procedure for determining the actual value of K [or C] exhibited by a meter under test.
Yet nowhere does the standard express (or even suggest) a degree of the deviation of the value of K [or C] from the stated value that is "permitted" under the standard.
So for some meter model, the manufacturer may state that it uses a value of 12.5 for K. In a laboratory, we determine by scrupulous and tedious testing the actual value of K practiced by a specimen meter and find that to be 11.2. Does that meter conform to the standard as to its accuracy? That is not defined.
Best regards,
Doug